<?xml version="1.0" encoding="utf-8"?>
<rss version="2.0" xmlns:atom="http://www.w3.org/2005/Atom">
	<channel>
		<title><![CDATA[SoliForum - 3D Printing Community — Double Shell]]></title>
		<link>https://www.soliforum.com/topic/13352/double-shell/</link>
		<atom:link href="https://www.soliforum.com/feed/rss/topic/13352/" rel="self" type="application/rss+xml" />
		<description><![CDATA[The most recent posts in Double Shell.]]></description>
		<lastBuildDate>Wed, 13 Jan 2016 20:56:51 +0000</lastBuildDate>
		<generator>PunBB</generator>
		<item>
			<title><![CDATA[Re: Double Shell]]></title>
			<link>https://www.soliforum.com/post/114067/#p114067</link>
			<description><![CDATA[<div class="quotebox"><cite>bobswika wrote:</cite><blockquote><p>You need to move the camera further up on the mounting post so that the cross hairs are aligned when the camera is perfectly square.&nbsp; The camera should not be tilted, it should be perfectly vertical.</p></blockquote></div><p>+1</p>]]></description>
			<author><![CDATA[null@example.com (pirvan)]]></author>
			<pubDate>Wed, 13 Jan 2016 20:56:51 +0000</pubDate>
			<guid>https://www.soliforum.com/post/114067/#p114067</guid>
		</item>
		<item>
			<title><![CDATA[Re: Double Shell]]></title>
			<link>https://www.soliforum.com/post/114066/#p114066</link>
			<description><![CDATA[<p>Thanks, I&#039;ll move the camera up and then make sure it is vertical.</p>]]></description>
			<author><![CDATA[null@example.com (kcassella)]]></author>
			<pubDate>Wed, 13 Jan 2016 20:53:43 +0000</pubDate>
			<guid>https://www.soliforum.com/post/114066/#p114066</guid>
		</item>
		<item>
			<title><![CDATA[Re: Double Shell]]></title>
			<link>https://www.soliforum.com/post/114055/#p114055</link>
			<description><![CDATA[<p>You need to move the camera further up on the mounting post so that the cross hairs are aligned when the camera is perfectly square.&nbsp; The camera should not be tilted, it should be perfectly vertical.</p>]]></description>
			<author><![CDATA[null@example.com (bobswika)]]></author>
			<pubDate>Wed, 13 Jan 2016 19:04:14 +0000</pubDate>
			<guid>https://www.soliforum.com/post/114055/#p114055</guid>
		</item>
		<item>
			<title><![CDATA[Re: Double Shell]]></title>
			<link>https://www.soliforum.com/post/114044/#p114044</link>
			<description><![CDATA[<p>pirvan</p><p>The main reason for asking, when I align the cross hairs to the tips of the calibration jig, the camera is tilted back and not square.&nbsp; Should I make the camera square and not worry about aligning to the top peaks of the jig?</p><p>Thanks</p>]]></description>
			<author><![CDATA[null@example.com (kcassella)]]></author>
			<pubDate>Wed, 13 Jan 2016 17:28:46 +0000</pubDate>
			<guid>https://www.soliforum.com/post/114044/#p114044</guid>
		</item>
		<item>
			<title><![CDATA[Re: Double Shell]]></title>
			<link>https://www.soliforum.com/post/113995/#p113995</link>
			<description><![CDATA[<p>There has been more than one calibration tool, so in the case of the original one with the 3 tips, then the horizontal crosshair lines up with the tips.&nbsp; </p><p>There&#039;s another calibration piece that uses a checkerboard pattern, so you line it up with the horizontal line edges of the squares, and I&#039;ve seen yet another 3D printed piece that seems to have some horizontal slots in it.</p><p>Whichever one you use, the principle is the same.</p>]]></description>
			<author><![CDATA[null@example.com (pirvan)]]></author>
			<pubDate>Wed, 13 Jan 2016 01:11:09 +0000</pubDate>
			<guid>https://www.soliforum.com/post/113995/#p113995</guid>
		</item>
		<item>
			<title><![CDATA[Re: Double Shell]]></title>
			<link>https://www.soliforum.com/post/113960/#p113960</link>
			<description><![CDATA[<p>Pirvan,</p><p>What are you saying &quot;Align the horizontal cross-hair with the horizontal line on the object&quot;.&nbsp; What horizontal line on the object.&nbsp; The instructions shows lining the horizontal cross-hair to the tips of the calibration jig.&nbsp; Here are the instructions I&#039;m using. <a href="http://www.murobo.com/start.html">http://www.murobo.com/start.html</a> <br />Is there a newer set of instructions?</p><p>Thanks for your help!!!</p>]]></description>
			<author><![CDATA[null@example.com (kcassella)]]></author>
			<pubDate>Tue, 12 Jan 2016 19:52:30 +0000</pubDate>
			<guid>https://www.soliforum.com/post/113960/#p113960</guid>
		</item>
		<item>
			<title><![CDATA[Re: Double Shell]]></title>
			<link>https://www.soliforum.com/post/113800/#p113800</link>
			<description><![CDATA[<p>Got it sorted out.&nbsp; After going through the calibration again the x axis measurements changed. <br />The one laser misses a little more than the other but I can live with that.</p><p><span class="postimg"><img src="http://soliforum.com/i/?OnbahN8.jpg" alt="http://soliforum.com/i/?OnbahN8.jpg" /></span></p>]]></description>
			<author><![CDATA[null@example.com (bobswika)]]></author>
			<pubDate>Sun, 10 Jan 2016 21:36:04 +0000</pubDate>
			<guid>https://www.soliforum.com/post/113800/#p113800</guid>
		</item>
		<item>
			<title><![CDATA[Re: Double Shell]]></title>
			<link>https://www.soliforum.com/post/113795/#p113795</link>
			<description><![CDATA[<p>I&#039;m sure what you say is correct, but boy is it hard to get it perfect.&nbsp; I went through the full calibration process again exactly as you described.&nbsp; It is much better but still the slightest bit off.&nbsp; Am I expecting too much?&nbsp; Do you think that the thicker fuzzy left laser line could be the problem.&nbsp; It&#039;s close to twice the thickness of the right.&nbsp; I attached pictures of my calibration and resulting test PNG, the lasers appear to be dead on.&nbsp; You can see the thicker laser on the right side of the pic.&nbsp; </p><p>Thanks a lot,</p><p>Bob</p><p><span class="postimg"><img src="http://soliforum.com/i/?MVNvkEw.jpg" alt="http://soliforum.com/i/?MVNvkEw.jpg" /></span></p><p><span class="postimg"><img src="http://soliforum.com/i/?Zv0vqCe.jpg" alt="http://soliforum.com/i/?Zv0vqCe.jpg" /></span></p>]]></description>
			<author><![CDATA[null@example.com (bobswika)]]></author>
			<pubDate>Sun, 10 Jan 2016 20:48:56 +0000</pubDate>
			<guid>https://www.soliforum.com/post/113795/#p113795</guid>
		</item>
		<item>
			<title><![CDATA[Re: Double Shell]]></title>
			<link>https://www.soliforum.com/post/113765/#p113765</link>
			<description><![CDATA[<p>If one shell starts to taper (inward or out), then the laser that produces it is not perfectly vertical or the camera itself is not properly aligned vertically.&nbsp; It may appear to be, but it&#039;s not.</p><p>I would suggest using a calibration object that is known to be perfectly vertical, and that has a slightly darker color.&nbsp; The reason for the darker color is that it will reduce the brightness and reflection of the laser on the surface and create a visibly &quot;thinner&quot; line. </p><p>I use a checkerboard pattern attached to a CD jewel box.&nbsp; When I open it up, the jewel box sits nice and vertical. It would also help if you find and mark the center of your turntable.</p><p>First, we need to line up the camera.&nbsp; Place your calibration object in the middle of the turn table with it&#039;s vertical line aligned with the centerline of the turntable.&nbsp; Next adjust the camera so it&#039;s vertical cross-hair lines up with the line on the calibration object.&nbsp; </p><p>Align the horizontal cross-hair with the horizontal line on the object. <strong>DO NOT TILT the camera</strong> to adjust it.&nbsp; The camera should always look straight, not up or down.&nbsp; You degree of tilt should be 0°.&nbsp; Use a square or protractor to make sure.</p><p>Once you have the camera aligned, tighten donw the screws just enough to prevent it from moving.&nbsp; Over tightening can lead to misalignment.</p><p>Now you have to align the lasers.&nbsp; first bring them as close to the center as possible so they intersect in the middle of the calibration tool, then calibrate each laser vertically one at the time by rotating them in their socket.&nbsp; Again, tighten the screws just enough to hold the laser, but don&#039;t over tighten.</p><p>When you have everything lined up, do an auto-calibration, and look at the resulting PNG file.&nbsp; Specifically, check the little green and yellow cross hairs at the top and bottom, they should ideally overlap, but if they&#039;re close, you should be good.</p><p>That&#039;s about all the information I can give you.&nbsp; If you do this right, your final output should have only one shell.</p><p>Good luck.</p>]]></description>
			<author><![CDATA[null@example.com (pirvan)]]></author>
			<pubDate>Sun, 10 Jan 2016 04:29:50 +0000</pubDate>
			<guid>https://www.soliforum.com/post/113765/#p113765</guid>
		</item>
		<item>
			<title><![CDATA[Double Shell]]></title>
			<link>https://www.soliforum.com/post/113761/#p113761</link>
			<description><![CDATA[<p>I have my scanner very close to being calibrated, I can&#039;t detect where it is off calibration if it is.&nbsp; I have used both the grid and the calibration object.&nbsp; I also used Pirvans suggestion to align the camera using an object on the end of the table, great idea.&nbsp; </p><p>I have one lingering issue that I&#039;m hoping someone can assist with.&nbsp; At the bottom of my scans the two lasers meet up perfect and there is no double shell at all.&nbsp; As the scan gets higher off from the table I start to get a double shell which gets consistently worse the higher up the object it goes. Does anyone have any suggestions what may be causing this.&nbsp; Thanks,</p><p>Bob</p>]]></description>
			<author><![CDATA[null@example.com (bobswika)]]></author>
			<pubDate>Sun, 10 Jan 2016 02:59:16 +0000</pubDate>
			<guid>https://www.soliforum.com/post/113761/#p113761</guid>
		</item>
	</channel>
</rss>
